This is a Multix error. Problem is in the Bucky stand, D30 fault maybe EPROM fault. DIL switch incorrectly set? CAN termination? Cable? Try to terminate CAN inside Polydoros.
There are too many devices in your system configuration that can generate that error number. Said error code must be preceded by another number that identifies the subsystem to which the error is associated.
It would be a format similar to this:
243/305
the first set of numbers identifies the device, the second the error.
I need that information to try to help you.
Gbenro Adekunle
Siemens polydoros error 305
March 15, 2022 11:40
Siemens polydoros error 305
to rate and post a comment
Tibor Ivanszky
re: Siemens polydoros error 305
March 17, 2022 10:13
This is a Multix error. Problem is in the Bucky stand, D30 fault maybe EPROM fault. DIL switch incorrectly set? CAN termination? Cable? Try to terminate CAN inside Polydoros.
to rate and post a comment
juan pablo amado
re: Siemens polydoros error 305
March 17, 2022 10:11
There are too many devices in your system configuration that can generate that error number. Said error code must be preceded by another number that identifies the subsystem to which the error is associated.
It would be a format similar to this:
243/305
the first set of numbers identifies the device, the second the error.
I need that information to try to help you.
to rate and post a comment
Germni
re: Siemens polydoros error 305
March 17, 2022 10:14
this error comes from TABLE ICONOS
you can check code on polydoros display D200 or code display D100
to rate and post a comment