Hello,
I am working with a Polydoros R/F 80. It is located at a low-income facility, hence all my work is pro bono.
Issue:
Upon start up, "ERR 0100" is displayed.
Board D470 was checked, and is good. However, voltage to board D470 from D810 is 270v, I was told it should be 565v.
Any help would be much appreciated!
Email: alchaar007@gmail.com
You should check the filament. Probably a damaged tube.
This is the error meaning.
Err 100 / Err 102 / Err 103 / Err 104: D470 Error
Component: Generator
Description: ⢠Error 100
Error during exposure, error of the Exposure Focus
⢠Error 102
Error during exposure
⢠Error 103
Error during exposure and Standby, Large Focus Error
⢠Error 104
Error during exposure and Standby, Small Focus Error
⢠D470 defective
Possible Error Causes:
⢠None; too high or too low heating current detected.
⢠Heating not learned if error occurred during exposure.
⢠Heating filament (focus) or high voltage cable broken.
⢠Cabling from and to the D470.
⢠D470 defective.
Special Reasons: (Error 102)
93:03 Small Focus
93:04 Large Focus
Corrective
Measures:
⢠Check the low voltages (Wiring Diagram).
⢠Check the ribbon cable from D800 to D470.X3.
⢠Check the small and large focus of the X-ray tube. Pre-heating takes place in standby by both focuses; if needed, replace the tube unit.
⢠During startup of the new system: The HV cables to the HV tankare switched.
⢠Check the HV cable (cathode side); if needed, replace the HV cable.
⢠Check the "VHF" voltage on the D470 board between X1, Pin 1 and Pin 3.
ô¹ Values: approx. 15 V, approx. 50 kHz
⢠Measure the intermediate circuit voltage on the D470 board between:
- Pins X4.3 and X4.6
ô¹ Basic: 565 V (±10%) in standby mode
ô¹ ESU: 460 V (±10%) in standby mode
- Pins X4.1 and X4.6
ô¹ 282 V (± 10%) in the standby mode
⢠Learn filament.
ô¹ See âSequence of the Tube Adjustment Stepsâ.
⢠On HVT/(D532), check the cable connections D532.X1 (large
focus) and D532.X3 (small focus).
⢠Replace the D470.
⢠Replace the tube.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
AHMAD ALCHAAR
Siemens Polydoros R/F 80 Error code 100
March 17, 2018 01:44
Hello,
I am working with a Polydoros R/F 80. It is located at a low-income facility, hence all my work is pro bono.
Issue:
Upon start up, "ERR 0100" is displayed.
Board D470 was checked, and is good. However, voltage to board D470 from D810 is 270v, I was told it should be 565v.
Any help would be much appreciated!
Email: alchaar007@gmail.com
to rate and post a comment
AHMAD ALCHAAR
re: Siemens Polydoros R/F 80 Error code 100
March 19, 2018 03:40
Anyone??
to rate and post a comment
Dante Nuevo
re: Siemens Polydoros R/F 80 Error code 100
March 22, 2018 09:29
You should check the filament. Probably a damaged tube.
This is the error meaning.
Err 100 / Err 102 / Err 103 / Err 104: D470 Error
Component: Generator
Description: ⢠Error 100
Error during exposure, error of the Exposure Focus
⢠Error 102
Error during exposure
⢠Error 103
Error during exposure and Standby, Large Focus Error
⢠Error 104
Error during exposure and Standby, Small Focus Error
⢠D470 defective
Possible Error Causes:
⢠None; too high or too low heating current detected.
⢠Heating not learned if error occurred during exposure.
⢠Heating filament (focus) or high voltage cable broken.
⢠Cabling from and to the D470.
⢠D470 defective.
Special Reasons: (Error 102)
93:03 Small Focus
93:04 Large Focus
Corrective
Measures:
⢠Check the low voltages (Wiring Diagram).
⢠Check the ribbon cable from D800 to D470.X3.
⢠Check the small and large focus of the X-ray tube. Pre-heating takes place in standby by both focuses; if needed, replace the tube unit.
⢠During startup of the new system: The HV cables to the HV tankare switched.
⢠Check the HV cable (cathode side); if needed, replace the HV cable.
⢠Check the "VHF" voltage on the D470 board between X1, Pin 1 and Pin 3.
ô¹ Values: approx. 15 V, approx. 50 kHz
⢠Measure the intermediate circuit voltage on the D470 board between:
- Pins X4.3 and X4.6
ô¹ Basic: 565 V (±10%) in standby mode
ô¹ ESU: 460 V (±10%) in standby mode
- Pins X4.1 and X4.6
ô¹ 282 V (± 10%) in the standby mode
⢠Learn filament.
ô¹ See âSequence of the Tube Adjustment Stepsâ.
⢠On HVT/(D532), check the cable connections D532.X1 (large
focus) and D532.X3 (small focus).
⢠Replace the D470.
⢠Replace the tube.
to rate and post a comment
gerardo millena
re: Siemens Polydoros R/F 80 Error code 100
May 07, 2018 09:29
Is it an intermittent error? If so you might have a ground-noise related. Try spray WD40 on CX connectors.