Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Siemens Polymobil Plus error code 39

(7)

(3)

Jorge Fernandez

Siemens Polymobil Plus error code 39

March 12, 2013 03:23

I would appreciate some help with a polymobil plus.
The unit will work normally while plugged in to ac Power. when I try to use it on battery Power it will give me an error code 39 and will not discharge.
Batteries have been replaced.
Thank you

Log inor Register

to rate and post a comment

(11)

OWALDO QUISPE PEREZ

re: Siemens Polymobil Plus error code 39

March 13, 2013 10:50

Check the frequency of the heating inverter at the D916 test
point FC1

Log inor Register

to rate and post a comment

(11)

OWALDO QUISPE PEREZ

re: Siemens Polymobil Plus error code 39

March 13, 2013 10:51

Check the frequency of the heating inverter at the D916 test
point FC1

Log inor Register

to rate and post a comment

(7)

(3)

Jorge Fernandez

re: re: Siemens Polymobil Plus error code 39

March 14, 2013 12:31

Thank you Osvaldo , I will Try this

Log inor Register

to rate and post a comment