Dear All,

Siemens Poly Mobile 3 we are using developed a fault and the transformer was changed.
After that we noticed an error 03 on the machine.
What can be the problem?
Thank you.

nestob2002@yahoo.com
+4368110738485

Best regards
Ernest Onuigbo

1 person liked this
Log inor Register

to rate and post a comment

Mohammed Abdullah

re: Siemens Polymobile 3

February 24, 2014 02:42

Dear Ernest,

Error 3 - Oil pressure rise or overheating of the tube

Error 3 Oil pressure rise or overheating of the tube
System status:
Standby
Explanation:
Action:
• Check the oil switch.
(Normal tube condition = switch closed)
􀂹 Warning: let the tube cool off before testing!
• Check the connections and the cable between the switch
and the D916 X8-9 and X8-8.

good luck

also i have all manuals for this system if you need them contact me

magmesme@gmail.com

regards

2 people liked this
Log inor Register

to rate and post a comment

(54)
(8)

Ernest ONUIGBO

re: re: Siemens Polymobile 3

February 25, 2014 08:51

Thank you Mohammed,I will send you a mail.
Meanwhile i got a tip from the guy using the machine that the Poly-mobile 3 is showing both error "03 & 06"
What could be the problem?

Log inor Register

to rate and post a comment

Mohammed Abdullah

re: re: re: Siemens Polymobile 3

February 25, 2014 12:15

Dear Ernesr,
Error 06 mean:
Error 6 - IH heating measured

Log inor Register

to rate and post a comment

Mohammed Abdullah

re: re: re: Siemens Polymobile 3

February 25, 2014 06:29

error 6 Explanation:
Action:
• Compare IH and IHS in D916 to make sure they are not the
same.
• Check the frequency of the heating inverter in the D916
TP - FC1.
• Check the form of the resonance current at the TP-I D916
(use Service Program 9).
􀂹 Possible causes are a bad cable connection from
D916 to D1(X8) (high resistance) to the single tank, a
defective heating inverter on board D922, a defective
heating transformer (single tank), the filament in the
single tank is defective, the D916 and/or the D922 are
defective, or the tube is defective.

Log inor Register

to rate and post a comment

(113)
(7)
(1)

jorge mazariegos

re: re: Siemens Polymobile 3

February 26, 2016 08:45

good afternoon sorry I look for the service manual siemens polymobile 3 could help thanks

Log inor Register

to rate and post a comment

DMEDS

re: re: re: Siemens Polymobile 3

February 28, 2016 04:16

Dear Sir,

I have the manuals you need,

if you need it contact me,

magmesme@gmail.com

Regards

Log inor Register

to rate and post a comment

(76)
(10)
(2)

hemrajsinh chauhan

re: re: re: re: Siemens Polymobile 3

March 20, 2019 09:56

I need manual and schematics pleas send me on raj_meditech@hotmail.com

Log inor Register

to rate and post a comment

(76)
(10)
(2)

hemrajsinh chauhan

re: re: Siemens Polymobile 3

March 20, 2019 09:56

I need it sir please send me on raj_meditech@hotmail.com

Log inor Register

to rate and post a comment

(56)
(2)
(8)

Jairo Manzaneda

re: Siemens Polymobile 3

February 25, 2014 08:43

Hi Ernest, is so important to have which polimobil 3 model do you have, because there are two version, one serial 3000 and up and another serial 3000 down and new models serial 2000 or higher, what is you model?....

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE Monitor Service You Can Trust

Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)