Siemens Polyphos 30M - display problems

Tags: Radiology - R/F, SIEMENS

(2)

soren larsen

Siemens Polyphos 30M - display problems

November 16, 2016 10:04

Hello
Siemens Polyphos 30M with iontomat (AEC), x-ray tube P125/30 C-1.3:
Display showing Error 14 => New similar tube recently installed.
Momentarily kV and mAs displays are not functioning - only sporadically they are lit.
According to technician, the display decoder PCB 4626727 D30 board is faulty or has a loose connection.
Any good ideas or comments would be received with gratitude!
Regards Soren

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: Siemens Polyphos 30M - display problems

November 18, 2016 10:19

the Error 14 is trouble on mAs tube, about board display only check the power supply +5 Vcc and all conection for this.
the fuse board have led on each voltajes.

Log inor Register

to rate and post a comment

(2)

soren larsen

re: re: Siemens Polyphos 30M - display problems

November 24, 2016 12:27

Thanks
Error 14 is fixed.
The displays showing kV and mAs on the generator are not working (sometimes no light, sometimes little, sometimes full light). Do you have an idea how to fix this problem?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)