Siemens RF system, model ICONOS R100 err code manual
April 06, 2012 12:31
Hello everyone, i need help regarding Siemens RF system, model ICONOS R100, Generator LX50, Err code 501,32, i need Err code Manual or solution what i do i am waiting for reply.
supremeditech@yahoo.com
Re: Siemens RF system, model ICONOS R100 err code manual
April 06, 2012 05:54
Error No. 501 of AP ID 32
Meaning:
CAN bus error
Reason:
- CAN transmit malfunction.
- No response to CAN read request.
- Lenze rectifier defective (see Additional Info).
Action:
- Check the 15 V voltage to D41 in the basic unit, LED_B12 ON.
- Check whether 240 V are present at the X1 frequency converter and whether the green LED is blinking/on.
- Check whether the terminal resistor is jumpered with JP5_1-2 on the D41 in the basic unit.
- Check the CAN bus cabling.
Additional Info Regarding Error Readout 501/32, Meaning of the Variable Values S1, S2
The defective rectifier can be found using the S1, S2 variable values.
- Prüfe 15 V-Spannung auf D41 des Grundgerätes, LED_B12 ON.
- Prüfen, ob am Frequenzumrichter an X1, 240 V anliegen und die grüne LED blinkt/leuchtet.
- Prüfen, ob der Abschlußwiderstand mit JP5_1-2 auf D41 des Grundgerätes gejumpert ist.
- Prüfe CAN-Bus-Verdrahtung.
Zusatzinfo zum ausgelesenen Error 501/32, Bedeutung der Variablen-Werte S1, S2
Über den Variablen-Wert von S1, S2 ist der fehlerhafte Umrichter zu finden.
Re: Siemens RF system, model ICONOS R100 err code manual
April 09, 2012 12:41
Hello Mr. Aftab
Did you sole you problem.?
Are you Iranian and where do you live?
If you have any help let me know. We can have more cooperation with each other.
Kazem Saremi
mk.saremi@gmail.com
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
sohail aftab
Siemens RF system, model ICONOS R100 err code manual
April 06, 2012 12:31
Hello everyone, i need help regarding Siemens RF system, model ICONOS R100, Generator LX50, Err code 501,32, i need Err code Manual or solution what i do i am waiting for reply.
supremeditech@yahoo.com
to rate and post a comment
Lenny Place
Re: Siemens RF system, model ICONOS R100 err code manual
April 06, 2012 05:54
Error No. 501 of AP ID 32
Meaning:
CAN bus error
Reason:
- CAN transmit malfunction.
- No response to CAN read request.
- Lenze rectifier defective (see Additional Info).
Action:
- Check the 15 V voltage to D41 in the basic unit, LED_B12 ON.
- Check whether 240 V are present at the X1 frequency converter and whether the green LED is blinking/on.
- Check whether the terminal resistor is jumpered with JP5_1-2 on the D41 in the basic unit.
- Check the CAN bus cabling.
Additional Info Regarding Error Readout 501/32, Meaning of the Variable Values S1, S2
The defective rectifier can be found using the S1, S2 variable values.
Variable Values for Units without Tomo
Variable Values for Units with Tomo
************************************************************************************************
Bedeutung:
Fehler CAN-Bus
Grund:
- CAN-Senden fehlerhaft.
- CAN-Leseanforderung nicht beantwortet.
- Lenze Umrichter defekt (siehe Zusatzinfo).
Aktion:
- Prüfe 15 V-Spannung auf D41 des Grundgerätes, LED_B12 ON.
- Prüfen, ob am Frequenzumrichter an X1, 240 V anliegen und die grüne LED blinkt/leuchtet.
- Prüfen, ob der Abschlußwiderstand mit JP5_1-2 auf D41 des Grundgerätes gejumpert ist.
- Prüfe CAN-Bus-Verdrahtung.
Zusatzinfo zum ausgelesenen Error 501/32, Bedeutung der Variablen-Werte S1, S2
Über den Variablen-Wert von S1, S2 ist der fehlerhafte Umrichter zu finden.
Variable-Werte bei Geräten ohne Schichten
Variable-Werte bei Geräten mit Schichten
Copyright ® 2000-2003 Siemens AG/SSW Help VA01C
to rate and post a comment
Kazem Saremi
Re: Siemens RF system, model ICONOS R100 err code manual
April 09, 2012 12:41
Hello Mr. Aftab
Did you sole you problem.?
Are you Iranian and where do you live?
If you have any help let me know. We can have more cooperation with each other.
Kazem Saremi
mk.saremi@gmail.com
to rate and post a comment