i FOUND FOLLOWING PROBLEM WITH SOMATOM AR STRR--
1.WHILE TAKING TOPOGRAM VERTICAL BANDS ARE APPEARING OVER THE IMAGE
2.WHILE TAKING TOMOGRAM-CONCENTRIC THICK RINGS APPEARING
IMges are not interpretable .kindly anybody give suggestions for the problem
thanking you
If you already done air calibration, and the ring persist, it's possible that, that one o some of the cams of the detector are bad.
You can try doing air calibration first, and see if they disappear.
Hope this helps.
Jorge Lopez
Have you try calibration? it seems like you might have bad detectors or a pcb on the system. process raw data to find out where in the system is the artifacts are coming from
Dear Sir:
Cable betwwen D3 &Datector should be checked(Black cable with special connector), D3 itdelf can be faulty.test trasmit with sramble,increase & test pattern.
With Bests
H-Shirazi
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Divi Srinivas
SIEMENS SOMATOM AR STAR PROBLEM
June 13, 2012 09:44
i FOUND FOLLOWING PROBLEM WITH SOMATOM AR STRR--
1.WHILE TAKING TOPOGRAM VERTICAL BANDS ARE APPEARING OVER THE IMAGE
2.WHILE TAKING TOMOGRAM-CONCENTRIC THICK RINGS APPEARING
IMges are not interpretable .kindly anybody give suggestions for the problem
thanking you
to rate and post a comment
JORGE LUIS LOPEZ
re: SIEMENS SOMATOM AR STAR PROBLEM
June 14, 2012 12:47
If you already done air calibration, and the ring persist, it's possible that, that one o some of the cams of the detector are bad.
You can try doing air calibration first, and see if they disappear.
Hope this helps.
Jorge Lopez
to rate and post a comment
Juan de la Cruz Martinez
re: SIEMENS SOMATOM AR STAR PROBLEM
June 13, 2012 11:43
Have you try calibration? it seems like you might have bad detectors or a pcb on the system. process raw data to find out where in the system is the artifacts are coming from
to rate and post a comment
Hossein Shirazi
re: SIEMENS SOMATOM AR STAR PROBLEM
June 13, 2012 11:41
Dear Sir:
Cable betwwen D3 &Datector should be checked(Black cable with special connector), D3 itdelf can be faulty.test trasmit with sramble,increase & test pattern.
With Bests
H-Shirazi
to rate and post a comment