Dear All.
My Somatom Plus 4 had a message while power up :"Generator power lost, please check fuses".
I checked all the fuses and also contact of the contactor are OK.
I checked 3 phase voltage on the DCC_CON_D650 X3 is present too.
Anybody had this experiences before? or anybody had idea what is wrong?.
I need also service manual for this machine.
Thanks in advanced for your kind help.
ronaldliem@yahoo.com
Hi Lenny Place, Could you send me a copy of Siemens Somatom Plus 4 Service manual? My e-mail address is f.panci@dns-medikal.al. Any kind of your help shall be much appreciated.
Best regards.
Are you sure ? Review all fuses of the left bottom in the generator cabinet. Remember in this system the 3 phases power from the main is divide in 2 ways. Can I will send to the service manual also. If you don't find the problem you can write me again
HI Francisco
This is ValentÃn Cardona
My Somatom Plus 4 expert dont go warm up, appear error MAS_700013; i dont have the error code list Can you help me
thanks
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Ronald Liem
Siemens Somatom Plus 4 problem
October 16, 2013 04:41
Dear All.
My Somatom Plus 4 had a message while power up :"Generator power lost, please check fuses".
I checked all the fuses and also contact of the contactor are OK.
I checked 3 phase voltage on the DCC_CON_D650 X3 is present too.
Anybody had this experiences before? or anybody had idea what is wrong?.
I need also service manual for this machine.
Thanks in advanced for your kind help.
ronaldliem@yahoo.com
to rate and post a comment
Lenny Place
re: Siemens Somatom Plus 4 problem
October 16, 2013 06:18
I will send the manual using google drive
to rate and post a comment
Vimal Prakash Tripathi
re: re: Siemens Somatom Plus 4 problem
October 17, 2013 07:26
dear lenny,
can i hve the service manual for ther somatom plus 4.
i will be very thankful to you.
regards
viaml
to rate and post a comment
Lenny Place
re: re: re: Siemens Somatom Plus 4 problem
October 17, 2013 09:24
you will be getting a email from Google Drive
to rate and post a comment
Ilhan Aktas
re: re: Siemens Somatom Plus 4 problem
November 04, 2013 09:40
Hello Lenny
send them to me in the service manual? error list in Have?
to rate and post a comment
Ronald
re: re: Siemens Somatom Plus 4 problem
March 10, 2014 08:31
I'm very sorry Lenny, This job take over by Dutch engineer,they are replaced D650 but the problem still persist, now back to me again.
to rate and post a comment
Fredi Panci
re: re: Siemens Somatom Plus 4 problem
March 26, 2014 08:13
Hi Lenny Place, Could you send me a copy of Siemens Somatom Plus 4 Service manual? My e-mail address is f.panci@dns-medikal.al. Any kind of your help shall be much appreciated.
Best regards.
to rate and post a comment
Lenny Place
re: re: re: Siemens Somatom Plus 4 problem
March 26, 2014 05:31
yes I will be sending it to you soon
to rate and post a comment
Rainner Lopez
re: re: Siemens Somatom Plus 4 problem
September 17, 2015 04:20
Hi lenny, can you send me the service manual pleases, my email is rainner.lopez@hotmail.com
to rate and post a comment
Rami Hadla
re: re: Siemens Somatom Plus 4 problem
September 22, 2015 08:19
Hello Lenny,
can you send this service for me:
ramihadlas@gmail.com
Thanks a lot,
Best Regards,
to rate and post a comment
Tomo Vul
re: re: Siemens Somatom Plus 4 problem
October 11, 2015 09:28
Hi lenny, can you send me the service manual pleases, my email is tomislavbre@gmail.com
to rate and post a comment
stark
re: re: re: Siemens Somatom Plus 4 problem
January 17, 2016 11:21
Hi Lenny ,can you send me the service manual please
my email is gaurav_76@yahoo.com
to rate and post a comment
Lenny Place
re: re: re: re: Siemens Somatom Plus 4 problem
January 17, 2016 05:30
Yes I will upload and send link to you
to rate and post a comment
Francisco Rojas
re: Siemens Somatom Plus 4 problem
October 16, 2013 11:08
Are you sure ? Review all fuses of the left bottom in the generator cabinet. Remember in this system the 3 phases power from the main is divide in 2 ways. Can I will send to the service manual also. If you don't find the problem you can write me again
to rate and post a comment
Vimal Prakash Tripathi
re: re: Siemens Somatom Plus 4 problem
October 17, 2013 07:26
can i hve the service manual for ther somatom plus 4.
i will be very thankful to you.
regards
vimal
to rate and post a comment
Valentin Cardona
re: re: Siemens Somatom Plus 4 problem
April 17, 2014 01:15
HI Francisco
This is ValentÃn Cardona
My Somatom Plus 4 expert dont go warm up, appear error MAS_700013; i dont have the error code list Can you help me
thanks
to rate and post a comment
Lenny Place
re: re: re: Siemens Somatom Plus 4 problem
April 17, 2014 01:39
can you perform calibrations? if not test the d640
to rate and post a comment
Valentin Cardona
re: re: re: Siemens Somatom Plus 4 problem
April 17, 2014 05:08
My mail valentinimagen@hotmail.com
to rate and post a comment
Francisco Rojas
re: re: re: Siemens Somatom Plus 4 problem
April 17, 2014 05:08
RAC problema: anode rotation is not OK. Check windings of anode should be 2-3 Ohms If it is bad change the tube if it is not change RAC
to rate and post a comment
Valentin Cardona
re: re: re: re: Siemens Somatom Plus 4 problem
April 17, 2014 06:02
Thanks Francisco, I check and let you what happen
to rate and post a comment
Francisco Rojas
re: Siemens Somatom Plus 4 problem
October 17, 2013 12:02
Have you the service manual now or not yet. Do you need more help?
to rate and post a comment
Vimal Prakash Tripathi
re: re: Siemens Somatom Plus 4 problem
November 02, 2013 09:11
dear can i hve the manual
my id is- anand_8050@yahoo.com
to rate and post a comment
Ronald
re: re: re: Siemens Somatom Plus 4 problem
March 10, 2014 10:43
Today, I found the problem is big contactor is stack,just repair and ok now, thanks everybody
to rate and post a comment
Ronald
re: re: re: Siemens Somatom Plus 4 problem
March 10, 2014 10:43
I had hard copy of the manual, I'm sorry unable to share with you
to rate and post a comment
Valentin Cardona
re: re: Siemens Somatom Plus 4 problem
April 17, 2014 05:08
Thanks Francisco, I have the service manual, but I dont have the error list
to rate and post a comment
Francisco Rojas
re: re: re: Siemens Somatom Plus 4 problem
April 18, 2014 09:03
In the software manual you have the interpration of errors.
to rate and post a comment