Siemens Symphony Maestro Class Problem

Tags: MRI, SIEMENS

(24)
(4)

WG

Siemens Symphony Maestro Class Problem

August 18, 2018 02:53

Looking for help with a difficult Siemens Symphony Maestro class fault.

The system is booting up fine and passing tuneup. However receive path calibration fails. I have tried two known working head coils, different ports on the table and a software reload. Still the same problem. Hardware test of the receive path pass in test tools.

Any thoughts on what may be the cause would be greatly appreciated.

Log inor Register

to rate and post a comment

(35)
(6)
(5)

Gary Provenzano

re: Siemens Symphony Maestro Class Problem

August 19, 2018 09:38

Look in session history, probable high noise, rf room?

Log inor Register

to rate and post a comment

(29)
(11)
(16)

Vijay Sharma

re: Siemens Symphony Maestro Class Problem

August 19, 2018 09:38

What error message are you getting while receive path calibration falling..

Log inor Register

to rate and post a comment