Signa Contour Gradient troubles

Tags: MRI

Posted by an Anonymous User

Signa Contour Gradient troubles

September 26, 2012 05:25

I got problems in a GE MRI Signa contour 3:
Gradient amplifier can not be recoverd
Any idea about this problem?
Thank you

Log inor Register

to rate and post a comment

(1)

(3)

Brijesh Suneja

re: Signa Contour Gradient troubles

September 27, 2012 07:15

pls give the error codes u are recieving, u will be getting some error on the gradient cabinet also.

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: Signa Contour Gradient troubles

September 27, 2012 12:21

Likely you have a bad gradient or a bad controller in the gradient. further information about the errors shown in the gradient cabinet and on the system screen are needed to help further.

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

re: re: Signa Contour Gradient troubles

September 27, 2012 01:56

I already check GPS, seems to be ok
but now about error I have on log:

Thu Sep 27 12:14:01 2012
Host: oc001 Proc: auto_film_composer Error:14160
File: main.c Line: 2021
Using AutoFilmComposer name

Thu Sep 27 12:14:44 2012
Host: oc001 Proc: SysMonitor Error:2240787
File: SysMonitor_func.c Line: 1154
The magnet room oxygen density is too low

Thu Sep 27 12:21:32 2012
Host: oc001 Proc: mrscan Error:2218642
File: ScanAct.cc Line: 3850
Scan aborted by System Monitor

************************************************************
Oxygen Alarm is ON, may this problem avoid scans?

Thank you

Log inor Register

to rate and post a comment

(37)
(6)

Dejun Li

re: Signa Contour Gradient troubles

September 27, 2012 08:22

First thing you need to do is "Reset TPS", or power off the system and reboot again.
What type of gradient amplifier do you have? 8280 or 8290?

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

re: re: Signa Contour Gradient troubles

September 28, 2012 08:24

thank you everybody, I already fix this problem. Adjusting O2 sensor, then oxygen alarm turned off automatically. after clear this error, system could get scans