Joel Grullon

silhouette service program

November 12, 2015 06:14

hello colleagues I need to know if the service software for ge silhouette x ray
(jedi generator )allow calibrate the filament, because I am triying it but only offer me the database for 2 tube types but I can't find another data base for a diferent x ray tube that not be included in this list. by that reason I need to calibrate. any idea I will apreciate

Log inor Register

to rate and post a comment

(6)

Jose Luis Baltasar

re: silhouette service program

November 14, 2015 04:59

Hello Joel
E-mal me I can help you jbaltasarf@gmail.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: silhouette service program

November 14, 2015 05:01

the filament aging (calibrations) are performed during the normal use of the system when exposure pulse widths exceed 10 ms. setup the system for 80KV 50mA and at least 20ms then perform 10 exposures perform for the filament. this will adjust the filaments.