SIREGRAPH CF - ERROR 10

(9)

(1)

THEOFILOS KIOUPIS

SIREGRAPH CF - ERROR 10

January 15, 2017 07:42

SWITCHING ON THE MACHINE COMES ERROR 10 , AND IS NOT POSSIBLE TO TAKE IN THE ADJUSTMENT MODE THE LIMITS 101 AND 102 ! AFTER PUTTING A NEW 3PT3 POT .
CAN ANYBODY PLEASE GIVE ME A HELP .
MY E-MAIL : theofilos.kioupis@yahoo.gr
thangs in advance !

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: SIREGRAPH CF - ERROR 10

January 17, 2017 10:25

Do you get the error when you first boot (Idle) when you try to move (Movement)?
Cassette
carriage
010 Idle Potentiometer 3PT3 defective
Potentiometer 3PT3 beyond end position
Cable connection broken / not connected
ADC input on 24S1 CPU board failed

010 Movement Potentiometer 3PT3 not mechanically connected
Potentiometer 3PT3 incorrectly wired
Motor 3PT3 incorrectly wired
Tacho generator 3DT3 failed / incorrectly wired

010 Controlled and no
movement
DAC output on 24S1 CPU board failed
Motor 3MT3 not connected
Board 23AZ2 has responded

Log inor Register

to rate and post a comment

(31)
(12)

Dionisis Avgoustidis

re: re: SIREGRAPH CF - ERROR 10

January 18, 2017 03:57

No move ?

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: SIREGRAPH CF - ERROR 10

January 19, 2017 09:04

So it boots up with the error message. Were you getting the error before you change the 3PT3 Pot?

Log inor Register

to rate and post a comment

(139)
(11)
(2)

ALEXANDER PILAFTSIS

re: re: re: re: SIREGRAPH CF - ERROR 10

February 06, 2019 08:41

Dear Lenny, We resieved error 098 on a similar -sieregraph CF - system. Do you have any idea what might be the problem? Can you pls send error list of this system?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: SIREGRAPH CF - ERROR 10

February 07, 2019 09:02

098 Console-control cabinet fiber optic cable defective
Problem in the initialization phase
Incorrect data transfer
please send me an email address I will send you the manuals

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)