Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

SIREMOBIL COMPACT error 4021

Tags: C-Arm

jawad habib

SIREMOBIL COMPACT error 4021

May 03, 2014 01:56

Dear Sirs,

I have a problem with:
SIREMOBIL COMPACT
error 4021 (no image)

please, what help can you give me?
can you sell me the service manual, and software
Type your question or comment here...

1 person liked this
Log inor Register

to rate and post a comment

Mohammed Abdullah

re: SIREMOBIL COMPACT error 4021

May 04, 2014 08:23

Dear Sir,

i have all manuals for this system,

if you need them contact me,

regards

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: SIREMOBIL COMPACT error 4021

May 04, 2014 08:24

Error E4021 = Memoskop does not answer
Possible trouble: Host repeats a task 3 times, error E4021 is then entered.
Possible help: Check connection cables to the Memoskop.
Possibly Memoskop or D1 defective..
------------------------------
In my experience an error 4021 is almost always a bad Memoskop,
email me your Memoskop type and we will give you a good price for a replacement.

Regards,

Wayne M. Horsman, VP
Columbia Imaging, Inc.
wayne@columbiaimaging.com

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: SIREMOBIL COMPACT error 4021

May 04, 2014 08:25

4021 Memoscope does not answer. D1 maybe faulty. Timekeeper chip battery expired? ivanski@spidernet.com.cy

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: re: SIREMOBIL COMPACT error 4021

May 05, 2014 10:22

1% D1, 99% Memoskop

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: re: SIREMOBIL COMPACT error 4021

May 05, 2014 10:22

99% Memoskop, 1% D1 board.