Dear friends, I need some help about the following errors in a Siremobil Compact (don't have the error list):4021----4024---2006---and 5014. Thankyou in advance. Claudio Balangione
Error E4021 Memoskop does not answer
Possible trouble: Host repeats a task 3 times, error E4021 is then entered.
Possible help: Check connection cables to the Memoskop.
Possibly Memoskop or D1 defective..
Error E4024 Memoskop initialization timeout
Possible trouble: Host registered a timeout of the Memoskop during the initialization of the system. Waiting time for the Memoskop reporting exceeded.
Possible help: Check Memoskop.
Check connection cable to the Memoskop.
Possibly Memoskop or D1 defective.
Error E2006 Memoskop, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Memoskop.
Possibly Memoskop or D1 defective.
Error F5014 Gen. Sirephos max oil pressure exceeded
Possible trouble: The oil pressure switch on the Sirephos has responded.
Possible help: Temperature of the Sirephos too high.
Check oil pressure switch.
Check cables to the oil pressure switch.
In the case of defective oil pressure switch, replace the Sirephos.
Monitoring of the oil pressure switch defective (board D1).
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
Claudio Balangione
Siremobil compact errors
February 13, 2012 05:11
Dear friends, I need some help about the following errors in a Siremobil Compact (don't have the error list):4021----4024---2006---and 5014. Thankyou in advance. Claudio Balangione
to rate and post a comment
Wayne Horsman
Re: Siremobil compact errors
February 17, 2012 09:46
Sounds like a bad memoskop.
Error E4021 Memoskop does not answer
Possible trouble: Host repeats a task 3 times, error E4021 is then entered.
Possible help: Check connection cables to the Memoskop.
Possibly Memoskop or D1 defective..
Error E4024 Memoskop initialization timeout
Possible trouble: Host registered a timeout of the Memoskop during the initialization of the system. Waiting time for the Memoskop reporting exceeded.
Possible help: Check Memoskop.
Check connection cable to the Memoskop.
Possibly Memoskop or D1 defective.
Error E2006 Memoskop, ser. receive line error
Possible trouble: The receive line has a constant level.
Possible help: Check receive line between D1 and Memoskop.
Possibly Memoskop or D1 defective.
Error F5014 Gen. Sirephos max oil pressure exceeded
Possible trouble: The oil pressure switch on the Sirephos has responded.
Possible help: Temperature of the Sirephos too high.
Check oil pressure switch.
Check cables to the oil pressure switch.
In the case of defective oil pressure switch, replace the Sirephos.
Monitoring of the oil pressure switch defective (board D1).
to rate and post a comment
Claudio Balangione
Re: Siremobil compact errors
February 20, 2012 10:16
Thank you again Waine. I will check those items. Claudio
to rate and post a comment
Gary Benitez
re: Re: Siremobil compact errors
June 05, 2019 08:49
Hi Claudio,
Did you fix your problem?
I am having the same issues except for one.
4021, 4024, 2006.
I don't have the 5014.
Please email me at gary@crownmedicalintl.com
Thanks in advance,
to rate and post a comment
Joao Carlso
re: Re: Siremobil compact errors
November 06, 2015 02:29
Dear friends, I need some help about the following error in a Siremobil Compact (don't have the error list): error 7150
Thanks
to rate and post a comment
dark mackerel
re: re: Re: Siremobil compact errors
November 13, 2015 04:18
7150 means the slot diaghragm rotation index position was not found during initialization