i am facing error0021. but i want to understand?
which board is ii power supply.
are 28 volts the high voltages that are out of tolerane,that error code also mentions?
28 volts are generated by D11, are these voltages going to ii via D6, if yes then what is the track?
Hi Fisal!
001E D11 +15V or -15V out of tolerance
Cause : Message from power supply board D11. The +15 V
. or -15V voltage is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective
. Connection between D11 and D6
. (acknowledgement)
.
.
.
001F D11, +24V out of tolerance
Cause : Message from power supply board D11. The +24 V
. voltage is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective
Connection between D11 and D6
(acknowledgement)
.
.
.
0020 D11, +28V out of tolerance
Cause : Message from power supply board D11.
. The +28 V voltage (operating voltage of I.I. power
. supply) is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective
. Connection between D11 and D6
. (acknowledgement)
.
.
.
0021 I.I. power supply error
Cause : Message from I.I. power supply. The power supply for
. the image intensifier (high voltage) is out-of-
. tolerance.
Active: Constantly
Possible trouble / help : I.I. power supply
. Interruption in indication lead from
. I.I. to board D6; operating voltage
. missing (also triggers error 0020).
.
.
.
0022 Area dose measuring device error
Cause : Message from board D16. Operating voltage
. for dose measuring chamber out-of-tolerance.
Active: Constantly
Possible trouble / help : Wrong configuration (measuring
. chamber selected, but not available)
. Measuring-chamber operating voltage
. missing. Board D16
You can also download sm and um from
www.gmecorp-usa.com/IM/XR/CA/SIEMENS/SV/SIRE/2000/siremobi
l2000.pdf and
www.amberusa.com/Manuals/Siemens_Siremobil_2000%20Users%20Manual.pdf
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
faisal
siremobile 2000 error0021
September 27, 2013 07:00
i am facing error0021. but i want to understand?
which board is ii power supply.
are 28 volts the high voltages that are out of tolerane,that error code also mentions?
28 volts are generated by D11, are these voltages going to ii via D6, if yes then what is the track?
to rate and post a comment
Marian Shvets
re: siremobile 2000 error0021
September 29, 2013 10:39
Hi Fisal!
001E D11 +15V or -15V out of tolerance
Cause : Message from power supply board D11. The +15 V
. or -15V voltage is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective
. Connection between D11 and D6
. (acknowledgement)
.
.
.
001F D11, +24V out of tolerance
Cause : Message from power supply board D11. The +24 V
. voltage is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective
Connection between D11 and D6
(acknowledgement)
.
.
.
0020 D11, +28V out of tolerance
Cause : Message from power supply board D11.
. The +28 V voltage (operating voltage of I.I. power
. supply) is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective
. Connection between D11 and D6
. (acknowledgement)
.
.
.
0021 I.I. power supply error
Cause : Message from I.I. power supply. The power supply for
. the image intensifier (high voltage) is out-of-
. tolerance.
Active: Constantly
Possible trouble / help : I.I. power supply
. Interruption in indication lead from
. I.I. to board D6; operating voltage
. missing (also triggers error 0020).
.
.
.
0022 Area dose measuring device error
Cause : Message from board D16. Operating voltage
. for dose measuring chamber out-of-tolerance.
Active: Constantly
Possible trouble / help : Wrong configuration (measuring
. chamber selected, but not available)
. Measuring-chamber operating voltage
. missing. Board D16
You can also download sm and um from
www.gmecorp-usa.com/IM/XR/CA/SIEMENS/SV/SIRE/2000/siremobi
l2000.pdf and
www.amberusa.com/Manuals/Siemens_Siremobil_2000%20Users%20Manual.pdf