SIRESKOP CX ERROR CODE 247 (32)

(9)
(2)

Menderes Dogan

SIRESKOP CX ERROR CODE 247 (32)

March 15, 2011 05:37

I need the meaning of error 247 FOR Sireskop cx. Thanking you. mendere_dogan@yahoo.com

Log inor Register

to rate and post a comment

(75)
(8)
(7)

JIFFY PAUL AKKARAKARAN

Re: SIRESKOP CX ERROR CODE 247 (32)

March 16, 2011 02:37

Hi Menderes Dogan!

The APID 32 means error is related to the sireskop CX and the error 247 means the error is in the collimator of the tube below the table. 247 says " Timeout during controlling height of the window". which means the window size height adjustment could not be performed in the specified time. Please check whether the vertical /height adjusting blades are moving when the machine initializes. If it is moving then check the 1 k multi turn potentiometer which works as an encoder for the collimator height measurement. The potentiometer must have gone bad so the system could not assess the position of the collimator blade position.

Good luck

Jiffy Paul Akkarakaran

Log inor Register

to rate and post a comment

(9)
(2)

Menderes Dogan

Re: SIRESKOP CX ERROR CODE 247 (32)

March 17, 2011 10:13

Thank you for help.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)