The APID 32 means error is related to the sireskop CX and the error 247 means the error is in the collimator of the tube below the table. 247 says " Timeout during controlling height of the window". which means the window size height adjustment could not be performed in the specified time. Please check whether the vertical /height adjusting blades are moving when the machine initializes. If it is moving then check the 1 k multi turn potentiometer which works as an encoder for the collimator height measurement. The potentiometer must have gone bad so the system could not assess the position of the collimator blade position.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Menderes Dogan
SIRESKOP CX ERROR CODE 247 (32)
March 15, 2011 05:37
I need the meaning of error 247 FOR Sireskop cx. Thanking you. mendere_dogan@yahoo.com
to rate and post a comment
JIFFY PAUL AKKARAKARAN
Re: SIRESKOP CX ERROR CODE 247 (32)
March 16, 2011 02:37
Hi Menderes Dogan!
The APID 32 means error is related to the sireskop CX and the error 247 means the error is in the collimator of the tube below the table. 247 says " Timeout during controlling height of the window". which means the window size height adjustment could not be performed in the specified time. Please check whether the vertical /height adjusting blades are moving when the machine initializes. If it is moving then check the 1 k multi turn potentiometer which works as an encoder for the collimator height measurement. The potentiometer must have gone bad so the system could not assess the position of the collimator blade position.
Good luck
Jiffy Paul Akkarakaran
to rate and post a comment
Menderes Dogan
Re: SIRESKOP CX ERROR CODE 247 (32)
March 17, 2011 10:13
Thank you for help.