Hi guys!
I have a problem with a table for siemens Sireskop CX ,There is a problem in Spotfilm device I need the values of resistors R3 .R4 Load position
- Park position - It is not found in catalog maintenance,,,
these are the 2 potentiometers? If so you need to follow the instructions on the service screen and perform the calibrations. Select the following operator interfaces in the service program:
- Adjustment
- Calibrations
- Spotfilm device
adel ebrahem
Sireskop CX table fix problem
November 18, 2012 11:07
Hi guys!
I have a problem with a table for siemens Sireskop CX ,There is a problem in Spotfilm device I need the values of resistors R3 .R4 Load position
- Park position - It is not found in catalog maintenance,,,
to rate and post a comment
sohail aftab
re: Sireskop CX table fix problem
November 19, 2012 05:28
please contact supermeditech@yahoo.com
to rate and post a comment
adel ebrahem
re: re: Sireskop CX table fix problem
November 27, 2012 07:44
So far we have not received one can anyone help me
to rate and post a comment
Lenny Place
re: Sireskop CX table fix problem
November 19, 2012 05:28
these are the 2 potentiometers? If so you need to follow the instructions on the service screen and perform the calibrations. Select the following operator interfaces in the service program:
- Adjustment
- Calibrations
- Spotfilm device
to rate and post a comment
adel ebrahem
re: re: Sireskop CX table fix problem
November 23, 2012 12:12
Needed for cD program and password . Possible to send to gentleman_22000@yahoo
to rate and post a comment
adel ebrahem
re: Sireskop CX table fix problem
November 23, 2012 12:12
i send to u
to rate and post a comment