Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Sireskop CX Table problem

Tags: Radiology - R/F

(73)
(8)

Kim Schofield

Sireskop CX Table problem

June 13, 2013 08:30

I have a Sireskop CX that will not tilt the table. Also the table top will only move a few inches before it stops and will not move again. At start up, Iam getting errors in the Kv/mA/mAs windows as follows:

mAs: 866
ms: 16

mAs: 258
ms: 32

I do not have service manual and need to locate one.

Thanks

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Sireskop CX Table problem

June 14, 2013 07:13

send me a email address, 32/258 could indicate a bad potentiometer or calibration the 16/866 is normally a videomed issue

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Sireskop CX Table problem

June 14, 2013 07:16

Does it have an overhead? ( 3D), if it does make sure it's on the park position.

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: re: Sireskop CX Table problem

June 15, 2013 07:29

Yes, it does have an overhead tube

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)