Sireskop sd service docs including table tilting problems

Tags: Radiology - R/F, SIEMENS

(5)

harry gandia

Sireskop sd service docs including table tilting problems

June 05, 2018 06:27

Hi dear i need service docs for sireskop sd andim facing the table tilting error 183/032 and 312/032
i Cant tilt Table i Have two errors permanent

ONE on start UP which is 312/032 =compression drive(KA): Wrong pos. of pot. during power on


and when i press optitouch to finish this err and then i press tilt button i see

err 183/032 = System tilt (SK): Motor controller Command Error

kindly tell me which is motor contoller board?


where is the problem in system?


thanks
Harry Gandia !
Free LancerDear Vesper i Cant tilt Table i Have two errors permanent

ONE on start UP which is 312/032 =compression drive(KA): Wrong pos. of pot. during power on


and when i press optitouch to finish this err and then i press tilt button i see

err 183/032 = System tilt (SK): Motor controller Command Error

kindly tell me which is motor contoller board?


where is the problem in system?


thanks
Harry Gandia !
Free Lancer

Log inor Register

to rate and post a comment

(9)
(1)

EFE87

re: Sireskop sd service docs including table tilting problems

June 07, 2018 02:57

You have some issue with the pot R116 (wiring, voltage, damage pot or adjustment) I will send the wiring diagrams for this Sireskop

Log inor Register