SOFTWARE INSTALLATION ERROR

Tags: Ultrasound

(2)

ochieng batista

SOFTWARE INSTALLATION ERROR

September 17, 2021 10:39

Output error file to the following location A:GHOSTERR.TXT

Log inor Register

to rate and post a comment

(50)
(5)

Ahmed Shawky

re: SOFTWARE INSTALLATION ERROR

September 20, 2021 10:19

the SW CD got corrupted at some point of ghost version installation, I recommend you to try with new SW CD. kindly tell me the model of device and software version u r trying to install and I can support u with SW and installation guide!

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: SOFTWARE INSTALLATION ERROR

September 20, 2021 10:20

This post is lacking a ton of missing information.
Are you trying to ghost a drive and reinstall that image on another drive?
Sounds like your Ghost Image might be corrupt.
Open the txt file using wordpad or notepad and see what the error is, it should guide you.
Instead of using ghost have you tried Macrium Reflect?
You may have to use a drive cradle to image the drive but it might work better than ghost.
Again, without any further information not sure how to answer you.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)