Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
hamayun kaleem
SOMATOM AR STAR
February 10, 2012 01:28
Hi,
i have problem with CT. the machine stop at warmup.
regards
to rate and post a comment
R.K.S singh
Re: SOMATOM AR STAR
February 10, 2012 04:22
please let me know the error
to rate and post a comment
hamayun kaleem
Re: SOMATOM AR STAR
February 12, 2012 11:31
Can you send me your email i will send you the screen shot of
error.
to rate and post a comment
James Fower
Re: SOMATOM AR STAR
February 10, 2012 11:45
The most common problem stop at warm up. Let me guess error " filament adaptation fail".
You can check the DC voltage to the inverter, filament board ( swing out rack) control board mounted on the gantry flat and vertical behind inverter board. When all checks good, then you will get to bad facts about the tube have a HV tank bad (the HV tank is inside the tube assembly).
Hopefully this was not the error to start wiyh. Good Luck
to rate and post a comment
hamayun kaleem
Re: SOMATOM AR STAR
February 11, 2012 11:45
Can you send me your email i will send you the screen shot of error.
to rate and post a comment
R.K.S singh
Re: SOMATOM AR STAR
February 12, 2012 06:56
123456roh@gmail.com
to rate and post a comment