April 06, 2017 05:32

Dear guys there is a Somaton Emotion 16 system showing: Reconstruction cancelled. Not all requested images available.
Before this, machine froze so operatos had to force a shutdown and a power on, then it only performs and shows scanograms but no further slices. Provides that system is warming up fine, I focus my attention on a software issue, and it is my belief that something is wrong with the patient database. Any help would be greatly appreciated.

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: Somatom Emotion 16

April 07, 2017 10:21

At first we need logs for be sure whats happened. Force shutdown operator console or IRS? You mean topogram scan is ok ?

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: Somatom Emotion 16

April 09, 2017 01:49

It looks your IRS has problem. please look at you Error Log, you will find more useful information.If your need service password, please email me: lindong123@hotmail.com