Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Somatom emotion 6

(182)
(31)

MIGUEL ANGEL TORRES MARTINEZ

Somatom emotion 6

August 21, 2016 10:59

I have a problem. FATAL RAC ERROR...emotion 6 (2003)

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Somatom emotion 6

August 22, 2016 10:14

Have you checked the fuses in the D400?

Log inor Register

to rate and post a comment

Eric Hauk

re: Somatom emotion 6

August 22, 2016 01:02

Fatal error (Rotate Anode Controller) check for power going to anode on the tube. If no power check generator/power box. If you have power replace the tube.