Somatom Volume Zoom Ring atifact

Tags: CT, SIEMENS

(44)
(4)

Rommel Saavedra

Somatom Volume Zoom Ring atifact

August 09, 2019 10:02

Hello
The images show ring artifact, I already did air calibration and the problem persists. I tried to make channel correction, but it shows the following message:
... calculating channel correction values failed, possibly caused by a defective channel...
What are the steps in the software to find and disable the defective channel?
Software version: Somaris/5 VA47C
I appreciate the help you can give me.

Log inor Register

to rate and post a comment

Eric Hauk

re: Somatom Volume Zoom Ring atifact

August 11, 2019 02:25

Likely bad DAS board, you need to troubleshoot further and figure out which one

Log inor Register

to rate and post a comment

Aaron Anderson

re: re: Somatom Volume Zoom Ring atifact

August 18, 2019 11:23

I believe he already made that conclusion Eric . I think hes not sure how to troubleshoot the specific detector board. Do you have any tips? I'm no Siemens expert.

Log inor Register

to rate and post a comment