Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

somaton plus 4

Tags: CT

(24)
(9)

cristian oliveira

somaton plus 4

February 28, 2014 10:08

good day all.

I'm working on a somaton plus 4 siemens, and some errors apears.
E 000014 resetwork timeout E0 0 14 1 f6 e fd
E 200022 inifret timeout 20 0 22 4 5a 0 0
E 600220
E 000011 initdone 0 0 11 10 96 00
E 100031


I need errors code documentation.

please help me



thanks

1 person liked this
Log inor Register

to rate and post a comment

(4)
(1)

Martin Delmonte

re: somaton plus 4

March 01, 2014 09:08

Hi cristian, I can help you

martin_delmonte@hotmail.com

1 person liked this
Log inor Register

to rate and post a comment

(47)
(15)

Francisco Rojas

re: somaton plus 4

March 01, 2014 09:09

Errors doesn´t clear. First error indicates that PTV is not OK , maybe D333 in the patient table is deffective. The last one D333 in GPC is not OK.

1 person liked this
Log inor Register

to rate and post a comment

(24)
(9)

cristian oliveira

re: re: somaton plus 4

March 01, 2014 08:06

This make some sense.
I had a problem on dc power on PHS and PTH oscilation them I changed the gpc whith pth and reloaded the firmware but the stop are opened, and I can not detect where is the stop.

Thanks

1 person liked this
Log inor Register

to rate and post a comment

(47)
(15)

Francisco Rojas

re: re: re: somaton plus 4

March 02, 2014 08:44

In the service manual there are one method for review the stop report loop

Log inor Register

to rate and post a comment

(24)
(9)

cristian oliveira

re: re: re: re: somaton plus 4

March 02, 2014 06:51

this is my problem. The manual I have is very limited. you could provide me the technical manual.

My email cristian.oliveiramedx@gmail.com

Thanks

1 person liked this
Log inor Register

to rate and post a comment

(47)
(15)

Francisco Rojas

re: re: re: re: re: somaton plus 4

March 03, 2014 08:57

Remember... plus 4 is it one old system from this the information is in CD. I can help to you by mail but I don´t to scanner all the information.

Log inor Register

to rate and post a comment

(47)
(15)

Francisco Rojas

re: re: re: re: re: somaton plus 4

March 03, 2014 08:57

Don,t forget the errors photo

Log inor Register