Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Sonoace 8000 live error code 85

Dawn Fowler

Sonoace 8000 live error code 85

January 07, 2015 10:11

I have a medison sonoace 8000 live. Last week all of the sudden it started throwing an error code 85 about 60 seconds into scanning in 2D and it freezes on error code screen and states it needs to reboot. Hopefully its an easy fix. Any advise would be greatly appreciated.

Log inor Register

to rate and post a comment

Tommy Lee

re: Sonoace 8000 live error code 85

January 07, 2015 05:47

Dear Dawn,
The Error Code 85 is meaning that the error concerned Motor Control Board.
Please disconnect the 3D probe and turn on the power switch and check again whether Error code is still appears or not.

Log inor Register

to rate and post a comment

Dawn Fowler

re: re: Sonoace 8000 live error code 85

January 08, 2015 08:50

I have disconnected the 3D probe and turned it back on. It has no code but I cannot do anything until I plug the probe back in. Everything works fine for about 60-70 seconds, then the error code appears.

Log inor Register

to rate and post a comment

Tommy Lee

re: re: re: Sonoace 8000 live error code 85

January 09, 2015 08:33

I think it is problem on the 3D/4D probe becuase when you turn on the power, system try to set the initial data. one of that is the set position crystal of 3D/4D probe - Crystal must be set on center position- but some time the positioning sensor in the probe was dead.
So, When you turn on the power, System request the initial data to each parts but probe did't set position to start and the Motor Contol board send wrong signal to CPU.
So, system show you Error message.
this Error message conerned Motor Control board and Probe but you told me during 60~70 seconds system was working well.
So, I doubt the Probe.

Log inor Register

to rate and post a comment

Dawn Fowler

re: re: re: re: Sonoace 8000 live error code 85

January 09, 2015 04:51

Yes, probe works fine in 2D right up till it hits the 60-70 second mark. I'm not sure if it functions well in 3D/4D because there's not really enough time for it to cycle and capture before it freezes on error code and requires reboot. I was hoping it is maybe just an update problem or some kind of setting that's off.

Log inor Register

to rate and post a comment

Tommy Lee

re: re: re: re: re: Sonoace 8000 live error code 85

January 09, 2015 06:49

When you press membrane of probe gently, you can feel the Crystal(it is hardest stuff) check that crystal is in center or not.
Normally, the positioning Sensor was Gone, the crystal is stuck in left or right side.
Please check that.

Log inor Register

to rate and post a comment

Dawn Fowler

re: re: re: re: re: re: Sonoace 8000 live error code 85

March 18, 2015 04:06

Sorry I just now saw that you had replied. I can not tell if it is in the center or not. But when I put it in 3D it does not move like it did before.

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: Sonoace 8000 live error code 85

January 07, 2015 05:48

Do you only have the 3D/4D probe? do you have a 2D probe? do you have a problem with any 2D probe? please contact me ricardo@ultrasolutions.com

Log inor Register

to rate and post a comment

Dawn Fowler

re: re: Sonoace 8000 live error code 85

January 08, 2015 08:50

I only have the 3D/4D probe. I have disconnected the 3D probe and turned it back on. It has no code but I cannot do anything until I plug the probe back in. Everything works fine for about 60-70 seconds, then the error code appears.

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: re: re: Sonoace 8000 live error code 85

January 08, 2015 04:00

Probably your probe bad, we can repair it, please contact me ricardo@ultrasolutions.com, thanks.

Log inor Register

to rate and post a comment

Dawn Fowler

re: re: re: re: Sonoace 8000 live error code 85

January 09, 2015 08:33

The probe only has no more than 75 hours on it. It shouldn't be bad.

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: re: re: re: re: Sonoace 8000 live error code 85

January 09, 2015 04:51

We can test the probe for you, we can do a an evaluation for free.

Log inor Register

to rate and post a comment

(8)

wilson cavalcante

re: Sonoace 8000 live error code 85

February 01, 2015 10:24

Friends, I need the schematics for the Low Voltage power supply for Medison SE, please contact me if you have it, my power supply burned a circuit and I can’t read the P/n of it, thanks. wcscavalcante@uol.com.br / wcscavalcante@ig.com.br

Log inor Register

to rate and post a comment

(1)

CUALLO

re: re: Sonoace 8000 live error code 85

July 07, 2025 12:00

Dear Wilson, did anyone help you with that problem ? I have a similar problem in PSU Module DMPS 3405 and need to know the part numbers (Q2) ZD2 and C5. Thank you

Log inor Register

to rate and post a comment

NGOUAMBE PAUL ARMAND

re: Sonoace 8000 live error code 85

August 21, 2017 08:30

dawn fowler i am, paul and am facing the same problem can you help with the way you resolve yours???

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Philips Achieva 1.5T MRI Spare Parts Available Looking for Philips Achieva 1.5T MRI Parts?

We offer reliable & cost-effective spare parts for the Philips Achieva 1.5T MRI system helping you minimize downtime & keep operations smooth. Wide inventory available. Quick turnaround time. Expert support. Worldwide delivery. Contact Limor@elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)