I have a problem with a sonoline elgra ultrasound.

It starts normaly but ther is no picture of a transducer. the rest of the picture is normaly. I have tested it with three diffrent transducers.

anyone an idea

lot of thanks

Pascal

Log inor Register

to rate and post a comment

(56)
(15)

imeng

re: sonoline elegra

December 26, 2013 06:43

Hi,
I dont understnd what you mean by no picture. but, if It is what I am thinking in, I think you have problem with -100V or +100V. so please check them in the back of the system. if one of them not there, check first of all the fuses on the right side near the big transformer.
if everything is ok, please back to us and we will think with you.