Hi everybody, I have an error EF7F0 in a Sonos 5500 and never start the application, I checked in the service manual and is a problem with DSCC board, I replaced it, but the problem persist, I reconect all the boards and put the original DSCC board and the error dissapear and work ok, but after 1 week, the problem appear again. will be the power supply the problem?, any glue?, someone can help me?
Regards
I do not think it would be a power supply problem.
Did you check the RST error log? It may provide more information. I would recommend running the Scanner subsystem diagnostics. The DSCC mainly communicates to the Beam boards.
Thanks to guide me I think the software is on mod or follpy disk and the procedure is as like other software installation
If there is change so pl guide me
Running the Scanner diagnostics, under the Test button, should isolate the failure. If not, you can try changing the position of the front 2 Beam boards with the 2 rear Beam boards. Your error was a DSCC 0 error. I think it will change to a DSCC 1 error if a Beam board is the problem. We have tested Beam and DSCC boards in stock.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Lobsang Ortiz
Sonos 5500 error EF7F0
January 19, 2015 05:08
Hi everybody, I have an error EF7F0 in a Sonos 5500 and never start the application, I checked in the service manual and is a problem with DSCC board, I replaced it, but the problem persist, I reconect all the boards and put the original DSCC board and the error dissapear and work ok, but after 1 week, the problem appear again. will be the power supply the problem?, any glue?, someone can help me?
Regards
to rate and post a comment
Jim Carr
re: Sonos 5500 error EF7F0
January 20, 2015 10:08
I do not think it would be a power supply problem.
Did you check the RST error log? It may provide more information. I would recommend running the Scanner subsystem diagnostics. The DSCC mainly communicates to the Beam boards.
to rate and post a comment
M.Y
re: re: Sonos 5500 error EF7F0
January 21, 2015 09:47
I have HP sonos 2000 and due cmos faulty system software loss so I need the system software for SONOS 2000
if you have pl help me at medilux@live.com
to rate and post a comment
EROTOKRITOS SKENTOS
re: re: re: Sonos 5500 error EF7F0
January 22, 2015 10:19
Do you know the procedure for loading the software because it is difficult.
to rate and post a comment
M.Y
re: re: re: re: Sonos 5500 error EF7F0
January 23, 2015 10:09
Thanks to guide me I think the software is on mod or follpy disk and the procedure is as like other software installation
If there is change so pl guide me
to rate and post a comment
EROTOKRITOS SKENTOS
re: re: re: re: re: Sonos 5500 error EF7F0
January 24, 2015 01:03
Let me check for the software and i will get back to you.What version do you need?
to rate and post a comment
M.Y
re: re: re: re: Sonos 5500 error EF7F0
January 23, 2015 10:10
medilux@live.com
to rate and post a comment
Marko Shevchuk
re: re: Sonos 5500 error EF7F0
January 23, 2015 10:11
Hello Jim!
Your information is very useful
and helped me!
to rate and post a comment
EROTOKRITOS SKENTOS
re: Sonos 5500 error EF7F0
January 21, 2015 09:49
Yo propably have a problem with one of the beam formers from the front.If you have one good board try replacing one by one.
to rate and post a comment
Lobsang Ortiz
re: re: Sonos 5500 error EF7F0
January 21, 2015 10:16
i don't have any beam former, but I try to get one and make the probes. Regards
to rate and post a comment
Jim Carr
re: re: re: Sonos 5500 error EF7F0
January 21, 2015 02:13
Running the Scanner diagnostics, under the Test button, should isolate the failure. If not, you can try changing the position of the front 2 Beam boards with the 2 rear Beam boards. Your error was a DSCC 0 error. I think it will change to a DSCC 1 error if a Beam board is the problem. We have tested Beam and DSCC boards in stock.
to rate and post a comment