GERARDO ALCANTARA

sonos 5500

January 03, 2013 03:49

Hi we have an Philips Sonos 5500 does not display anything on the touch screens and monitor,when powered on. All the fans seem to be working. The power supply module and the PC are confirmed working (not sure of the interface/PCI card on the PC) One other section I looked at wasthe proc. graphic in the card rack, Which displays DFD28-0002.

We check to fuses on yhe main I/O board under the keyboardand these components are not open, we check the power supply and the voltages are correctly.

can you help me with this issue.

thanks in advance

Log inor Register

to rate and post a comment

(43)
(3)
(4)

Rodolfo Reyes Ingeniero

re: sonos 5500

January 03, 2013 05:00

when disssembly the control panel inside check the cables that all are well connect

Log inor Register

to rate and post a comment

Aizaz Ali Malik

re: sonos 5500

January 04, 2013 10:14

Dear i think u shall check the Grpahic processor board that have fault confirmly

1 person liked this
Log inor Register

to rate and post a comment

(7)

Pir Muhammad ilyas

re: re: sonos 5500

August 24, 2015 04:28

How can the fault be
rectify?

Log inor Register

to rate and post a comment

Aizaz Ali Malik

re: sonos 5500

January 04, 2013 10:14

u have problem in digital graphic processor baord

1 person liked this
Log inor Register

to rate and post a comment

(53)
(10)

hiortsam dade

re: sonos 5500

August 25, 2015 04:10

The problem is with the digital graphic processor board

Log inor Register