Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Sonos 7500 - Error messages when storing files

Tags: Cardiac Imaging

(10)
(2)

Sharon Pyc

Sonos 7500 - Error messages when storing files

April 10, 2013 10:48

"Network interface error during study creation..." repeatedly comes up after acquiring each image during a study. Does anyone know how to stop this?
Also another error message "Sonos /IDI options are mismatched. A network storage malfunction will result ..." Comes up when trying to end a stress test.

1 person liked this
Log inor Register

to rate and post a comment

(35)
(2)
(3)

Robert Celmer

re: Sonos 7500 - Error messages when storing files

April 10, 2013 05:45

Email me..parts@nwmedicalequipment.com

Log inor Register

to rate and post a comment

(1)

Jet678

re: re: Sonos 7500 - Error messages when storing files

April 11, 2013 09:20

Reload solfware on idi

Log inor Register

to rate and post a comment

(1)

Frankie Robinson

re: Sonos 7500 - Error messages when storing files

April 11, 2013 09:19

Reload solfware on idi

Log inor Register

to rate and post a comment

(77)
(18)

Brian Eberling

re: Sonos 7500 - Error messages when storing files

April 11, 2013 09:20

Email me at beberling@waveimaging.net and we can get you all fixed up.

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: Sonos 7500 - Error messages when storing files

April 11, 2013 09:21

Did you replace a board, it looks like you lost all your configuration, that is why is giving you an error Network and the Sonos/IDI error, you lost configuration and Options, check battery on Processor Graphics board.

Log inor Register

to rate and post a comment

Bill Decker

re: Sonos 7500 - Error messages when storing files

April 11, 2013 09:27

Please contact Bill Decker at. randlultrasoundservice@wowway.com.

Log inor Register

to rate and post a comment

Bill Decker

re: Sonos 7500 - Error messages when storing files

April 11, 2013 12:02

It looks like you have the network format on the SonosPC setup for DICOM and the 7500 thinks its DSR or vice versa.

Or Network Interface, DICOM Store, or IDI option is turned off.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Cyclic Materials - Sell us your permanent magnet MRIs! WE BUY WARM MAGNETS FOR CASH

Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621

stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)