Sonosite 180 Error Problem

(10)
(1)

Sohail Butt

Sonosite 180 Error Problem

June 11, 2010 10:17

I am facing a problem in subject unit the fault detail is this when we do power on this unit unit start normal all display comes normal and when Image view also start normal then sudden Image area go black and these error comes on the cornal or display screen Error detail is this C.0X000004FC(1276)
D.0X00060000(393216)
This procedure happen only in less then one minute after start and then no function work like power off also doesn't work
if some body know and can help please contact.

My E.mail sohailbutt12@yahoo.com

Log inor Register

to rate and post a comment

Alessandro Clarizia

Re: Sonosite 180 Error Problem

October 27, 2010 06:18

Hi Sohail
try to clean the probe connector (from ultrasound and probe side)
Good luck !