Hello ,
I need the service manuals for Spacelabs 90303B & Spacelabs 90309,
So if you please send it to me on e-mail samirmedical@hotmail.com and thanks to you.
could you please send me the service manual of the SpaceLabs 90309
or,I have this monitor without the power supply, do any one knows how to connect the 24v directly to the J1, I have another power supply
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
Luis Togo
Spacelabs 90303B Problems
August 04, 2009 03:28
Hello:
I have a Spacelabs 90303 which display doesn't switch ON. The module switches correctly the status LEDs are flashing random (as it supposes to be).
I've checked the power supply and I found that the +58V used for the display signals is not working properly.
I get a 250ms pulse every second of about 15V instead of the +58V.Also the PFAIL signal is generated.
I assume there might be kind of a shortcircuit or a damaged ELKO.
Does anybody has an idea where should I look(on which components?
I have the service manuals but not the schematics of the power supply.
Thanks!
to rate and post a comment
Mohmmad Samir Hawout
Re: Spacelabs 90303B Problems
June 30, 2010 05:09
Hello ,
I need the service manuals for Spacelabs 90303B & Spacelabs 90309,
So if you please send it to me on e-mail samirmedical@hotmail.com and thanks to you.
to rate and post a comment
Abdulhay Khatib
re: Re: Spacelabs 90303B Problems
May 06, 2014 11:53
could you please send me the service manual of the SpaceLabs 90309
or,I have this monitor without the power supply, do any one knows how to connect the 24v directly to the J1, I have another power supply
to rate and post a comment