Over 500 Total Lots Up For Auction at One Location - NJ 06/15

SRU problem on LightSpeed QX/i CT

Tags: CT, GE

(59)
(2)

Abu-Yusuf

SRU problem on LightSpeed QX/i CT

August 27, 2014 10:08

Hi all, i have met a problem in CT LightSpeed. While machine is standing on idle mode after a several hours it gives an error saying that "SRU has gone down. Restart your system to recover" Does anybody knows whats wrong with it.
Thanks

Log inor Register

to rate and post a comment

(62)
(11)

zekeng brice

re: SRU problem on LightSpeed QX/i CT

September 02, 2014 04:36

after sru has gone down what are exactly the errors messages at the errors log ?did the scan ok reconstructions ok before gone down?
The err log may help to determine the fault it may be the ig ;the communication issue or something else

Log inor Register

to rate and post a comment