STARTUP PROBLEM ON OEC9000 C-ARM

(25)
(6)

Eduardo Vega

STARTUP PROBLEM ON OEC9000 C-ARM

June 26, 2006 08:00

HELLO!!!! I AM HAVING A PROBLEM WITH THE STARTUP SEQUENCE ON A C-ARM OEC9000, WHEN I TURN ON THE MACHINE THE MESSAGE ON C-ARM DISPLAY STAY IN "PLASE WAIT B8" AND NO GO FURTHER, IF ANYONE HAVE A COMMENT WILL BE APPRECIATE.

Log inor Register

to rate and post a comment

(93)
(7)

Edward Yan

Re: STARTUP PROBLEM ON OEC9000 C-ARM

July 03, 2006 01:45

B8 = Initializing stack pointer. B8 boot code means you were stuck at booting Technique Processor board. Maybe you had a faulty Technique Processor board or a corrupted floppy disk. Double check the Technique Processor's LED read out.

Enter control panel status mode for further troubleshooting.

www.freedownloadtools.com has "PC bug doctor" free trial.

Edward Yan, B.S.E.E.

Log inor Register

to rate and post a comment

(25)
(6)

Eduardo Vega

Re: STARTUP PROBLEM ON OEC9000 C-ARM

July 03, 2006 04:56

Yes, in the Technique Pocessor Bd. i have an 8 in the display, and i also tried with two differents floppy disk with the same result; is possible that may i have a problem with the floppy drive?
Thank you so much for your comment.!!!!!!!!

Log inor Register

to rate and post a comment

(93)
(7)

Edward Yan

Re: STARTUP PROBLEM ON OEC9000 C-ARM

July 05, 2006 08:30

You have also error 8 in the LED display at the Technique Processor board it means you have a faulty TP board. Are the ICs in the sockets? Have you tried pushing the ICs into the sockets trick with power off? Inspect the board components and copper lands. Check the on board battery. You do not have a floppy drive problem because it reads the floppy disks. You do not have floppy disks problem because you tried two different disks already.

Log inor Register

to rate and post a comment

(25)
(6)

Eduardo Vega

Re: STARTUP PROBLEM ON OEC9000 C-ARM

July 05, 2006 08:30

Thank you for the advise i will do that with the TP board and i hope that it works!!!!
Best Regards!!!!!!!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)