Hello,I have a problem with Statim 5000.It's vibrator pump produces intense sound and vibration and it doesn't pull the water in.I cleaned up the pump and everything was ok .I measure with Ohms and it displays 68 KΩ.Is this result normal?I measure the volts on the vibrator while its active and i get 275 Volt.Where do you think is the problem ,on then vibrator pump or the CPU ?The vibrator pump which is on the machine works with 220 volts.All suggestions are appreciated.Kind Regards
Good afternoon:
also I have a problem with a model 2000S STATIM I send the error is nr 60 and not indicating any suggestions please or documentation.
regards
Oswaldo
Dont loose time with the pump. You have proble in non return valve, praticly the small ball at the end of pump.Replace it. But buy ceme, not ulke. It is the same from coffe machines, price aboute 20€
Good afternoon:
also I have a problem with a model 2000S STATIM I send the error is nr 60 and not indicating any suggestions please or documentation.
regards
Oswaldo
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
Gennadios Tsaptsidis
Statim 5000 Problem with Vibrator Pump
October 12, 2013 09:11
Hello,I have a problem with Statim 5000.It's vibrator pump produces intense sound and vibration and it doesn't pull the water in.I cleaned up the pump and everything was ok .I measure with Ohms and it displays 68 KΩ.Is this result normal?I measure the volts on the vibrator while its active and i get 275 Volt.Where do you think is the problem ,on then vibrator pump or the CPU ?The vibrator pump which is on the machine works with 220 volts.All suggestions are appreciated.Kind Regards
to rate and post a comment
alexium
re: Statim 5000 Problem with Vibrator Pump
October 13, 2013 06:31
Hello Gennadios! Read this article:
http://www.home-barista.com/espresso-machines/repairing-ulka-vibratory-pump-t1079.html
http://coffeetime.wikidot.com/vibration-pump-is-it-worn-out
to rate and post a comment
Matt Du Vall
re: Statim 5000 Problem with Vibrator Pump
October 14, 2013 07:28
I had a simular problem. I replaced the air filter and the problem went away.
to rate and post a comment
OWALDO QUISPE PEREZ
re: re: Statim 5000 Problem with Vibrator Pump
October 14, 2013 04:57
Good afternoon:
also I have a problem with a model 2000S STATIM I send the error is nr 60 and not indicating any suggestions please or documentation.
regards
Oswaldo
to rate and post a comment
bojan berlot
re: Statim 5000 Problem with Vibrator Pump
October 14, 2013 07:35
Dont loose time with the pump. You have proble in non return valve, praticly the small ball at the end of pump.Replace it. But buy ceme, not ulke. It is the same from coffe machines, price aboute 20€
to rate and post a comment
OWALDO QUISPE PEREZ
re: re: Statim 5000 Problem with Vibrator Pump
October 14, 2013 04:57
Good afternoon:
also I have a problem with a model 2000S STATIM I send the error is nr 60 and not indicating any suggestions please or documentation.
regards
Oswaldo
to rate and post a comment
bojan berlot
re: re: re: Statim 5000 Problem with Vibrator Pump
October 15, 2013 05:56
Cycle Fault #60 For a 134°C cycle, the chamber temperature dropped below the sterilization temperature, allowing for measurement error (i.e. Tchm
to rate and post a comment