Colin,
It sounds like the digital signal board or hard drive has failed in the monitor cart. I service several of these an I have all the manuals. Parts are hard to come by for this machine.
Colin Kelley Jr.
Stenoscop 2 boot errors
February 24, 2011 09:53
We are working on a 1992 Stenoscop which gives the following errors on boot up:
DMSP SEQUENTIAL BUS ES FAILED
DSAD DIGITALIZATION FAILED
DSMP OPERATIONS TEST FAILED
POWER UP SELF TEST FAILED
Please advise starting point as we have no experience working on this unit.
Thanks-
Colin@kelleyxray.com
to rate and post a comment
Deon Armes
Re: Stenoscop 2 boot errors
March 04, 2011 09:52
Colin,
It sounds like the digital signal board or hard drive has failed in the monitor cart. I service several of these an I have all the manuals. Parts are hard to come by for this machine.
Deon
to rate and post a comment
nasir khan
re: Re: Stenoscop 2 boot errors
January 01, 2013 02:18
i have the same problem with the HDD which changed but now i need the software can you please help me in that ????