Re: Sterling/dupont lp 300/400 Initializing Problem
September 30, 2006 02:17
If you are hearing periodic clicks from laser engine-try to adjust EOS/SOS signals. if not, measure +5 on the backplane behind of the imager. if it is out of tollerance(the value must be 5.0-5.25), adjust +5 volts by variable resistor inside the power supply assy.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Blanca Parciak
Sterling/dupont lp 300/400 Initializing Problem
September 29, 2006 02:40
Hi I have a problem with this laser imager, everything looks to be ok but it stay in code 00 "initializing" any help would be appreciate
to rate and post a comment
Kakhaber Gvaramia
Re: Sterling/dupont lp 300/400 Initializing Problem
September 30, 2006 02:17
If you are hearing periodic clicks from laser engine-try to adjust EOS/SOS signals. if not, measure +5 on the backplane behind of the imager. if it is out of tollerance(the value must be 5.0-5.25), adjust +5 volts by variable resistor inside the power supply assy.