Sterling/dupont lp 300/400 Initializing Problem

Blanca Parciak

Sterling/dupont lp 300/400 Initializing Problem

September 29, 2006 02:40

Hi I have a problem with this laser imager, everything looks to be ok but it stay in code 00 "initializing" any help would be appreciate

Log inor Register

to rate and post a comment

(49)

Kakhaber Gvaramia

Re: Sterling/dupont lp 300/400 Initializing Problem

September 30, 2006 02:17

If you are hearing periodic clicks from laser engine-try to adjust EOS/SOS signals. if not, measure +5 on the backplane behind of the imager. if it is out of tollerance(the value must be 5.0-5.25), adjust +5 volts by variable resistor inside the power supply assy.

Log inor Register

to rate and post a comment