Dear friends,
I have an inconvenient with x-ray equipment brand SWISSRAY model DDR COMBI PLUS.
The error message is:
"W1 42_010: The system is not x-ray and/or offset calibrated. The image quality may be reduced. Please, contact the support center for more information"

Log inor Register

to rate and post a comment

(14)
(2)
(2)

Aaron Bassham

re:SWISSRAY DDR COMBI PLUS

May 29, 2018 03:55

The detector needs calibration. Recommended to always make a copy of all references before renewing the calibration. Perform detector calibration starting with Offset Calibration.