check cabling at collimator encoder andsurrounding area..check that collimator encoder light is green/lit...check config of system to ensure that all subsystems show as available and not si,ulation..check all CAN nodes are indicating as grren in system config..if all good then recal Stand
Hi,how are you ?
* E5 75_102 collimator sensor does not exist
* E5 57_102 Bucky sensor does not exist
* E5 70_125 if the Tubedrive is correct calibrated, press F3 on remote control
* E5 70_102 Tubedrive does not exist
We have found the above problems, thanks for your help.
But, we have the problem now, the device is giving the E2 20_105 "IPS not ready" error.
Is there a problem in SwissVision program?
can you help me?
for some reason the Swissvision IPS is not communicating with the Expert..it could be a dcom issue btwn the two..it could be cameras not ready scenario with cameras not cooling down to minus 10 degrees due to perhaps a tubing issue..was the system working okay then stopped working..is there maybe a n ips nic card issue and the it settings are incorrect now or needs to be reset. is the swissvision database full and you need to delete images..can you ping the ips from the expert.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Arnie Mataro
SWISSRAY DRF
February 08, 2017 10:18
Anyone can give a service manual, installation and start up instruction for Swissray x ray ddr element?
Many Thanks.
to rate and post a comment
Lenny Place
re: SWISSRAY DRF
February 09, 2017 08:45
I can send you some manuals on it that might help/ need an email address
to rate and post a comment
Alexis
re: SWISSRAY DRF
February 09, 2017 08:49
Hi Arnie, I have all manuals pls contact me at tecnoray@gmail.com
to rate and post a comment
Semih Gunduz
re: re: SWISSRAY DRF
March 04, 2018 10:17
Hello Alexis,
Can you please send me swissray manuals you have?
We have some problem with a system installation.
* E5 75_102 collimator sensor does not exist
* E5 57_102 Bucky sensor does not exist
* E5 70_125 if the Tubedrive is correct calibrated,press F3 on remote control
* E5 70_102 Tubedrive does not exist
Thanks alot
to rate and post a comment
rsimaging
re: re: re: SWISSRAY DRF
March 05, 2018 10:32
check cabling at collimator encoder andsurrounding area..check that collimator encoder light is green/lit...check config of system to ensure that all subsystems show as available and not si,ulation..check all CAN nodes are indicating as grren in system config..if all good then recal Stand
to rate and post a comment
Ercument Ergur
re: re: re: re: SWISSRAY DRF
March 13, 2018 09:07
Hi,how are you ?
* E5 75_102 collimator sensor does not exist
* E5 57_102 Bucky sensor does not exist
* E5 70_125 if the Tubedrive is correct calibrated, press F3 on remote control
* E5 70_102 Tubedrive does not exist
We have found the above problems, thanks for your help.
But, we have the problem now, the device is giving the E2 20_105 "IPS not ready" error.
Is there a problem in SwissVision program?
can you help me?
to rate and post a comment
rsimaging
re: re: re: re: re: SWISSRAY DRF
March 13, 2018 01:21
curious what your fix was for the other issues..again please what model of Swissray do you have. do you have any other error codes.
to rate and post a comment
Ercument Ergur
re: re: re: re: re: re: SWISSRAY DRF
March 14, 2018 09:40
Swissray ddRModulaire plus ( year 2004 )
eXpert 4000
GEN-X-4000
Workstation 670
ERROR only , E2 20_105 '' IPS not ready ''
no other error codes.
to rate and post a comment
rsimaging
re: re: re: re: re: re: re: SWISSRAY DRF
March 14, 2018 03:12
for some reason the Swissvision IPS is not communicating with the Expert..it could be a dcom issue btwn the two..it could be cameras not ready scenario with cameras not cooling down to minus 10 degrees due to perhaps a tubing issue..was the system working okay then stopped working..is there maybe a n ips nic card issue and the it settings are incorrect now or needs to be reset. is the swissvision database full and you need to delete images..can you ping the ips from the expert.