Am working on a GE Revolution Maxima that is aborting scan it gives the following error message:
1. Rail temperature out of tolerance
Sensor: 7
Temperature: 25.7
Limits: 37. 39.
position ID: 6-left 7-center 8-right
2. Detector Module Temperature out of tolerance
Detector Module: 1
Detector module temperature: 26.69
Limits: 37 42
what could be wrong? please help its urgent
Looks like your rail temperature is low.
Could be an issue with the rail or modules.
You have 6 detector modules and the system is reporting module #1 as out of tolerance....its temperature is low
Contact me to troubleshoot this
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Chidi Agunonu
system aborting scan gives error message
June 02, 2025 06:01
Am working on a GE Revolution Maxima that is aborting scan it gives the following error message:
1. Rail temperature out of tolerance
Sensor: 7
Temperature: 25.7
Limits: 37. 39.
position ID: 6-left 7-center 8-right
2. Detector Module Temperature out of tolerance
Detector Module: 1
Detector module temperature: 26.69
Limits: 37 42
what could be wrong? please help its urgent
to rate and post a comment
CRAIG WAMBURA
re: system aborting scan gives error message
June 06, 2025 09:54
Looks like your rail temperature is low.
Could be an issue with the rail or modules.
You have 6 detector modules and the system is reporting module #1 as out of tolerance....its temperature is low
Contact me to troubleshoot this
to rate and post a comment
Chidi Agunonu
re: re: system aborting scan gives error message
June 10, 2025 10:12
so why is the temperature going low does it mean the heater is not working?