juan pablo amado

system five

May 08, 2018 11:53

Startup problems, after replacement of NVRAM
The factory settings were loaded, but I do not have the memory size configuration data. I need the address to correct to define ram size for cpu MVME167. now shows 0x4000000 and need define 0x2000000 BSP

Log inor Register

to rate and post a comment

(4)

Jozsef Sipeki

re: system five

May 10, 2018 09:22

The NVRAM full reprograming not possible in the System Five. Before use a new NVRAM You can load factory default data outside from CPU MVME167. I could help you! My E-mail: sipeki@medin.hu

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: system five

May 15, 2018 09:54

I just wrote to your dotmed msg. box, your mail rejects msg. from me. My mail is vitalmed@email.com