sytec 3000i aperture error

(1)

Ivonne Cardona

sytec 3000i aperture error

September 17, 2005 11:24

I have an aperture error, the collimator aperture is not working o.k., when I do a test with tgp board I have a error. Can anyone tell me what can I do???
thank you

Log inor Register

to rate and post a comment

(58)
(6)

Elias Valdes

Re: sytec 3000i aperture error

September 19, 2005 02:32

Make sure the motor tries to move, sometimes the screws that are on the motor shaft become loose causing an error and not allowing the sensors to register the asked collimator apperture.

Log inor Register

to rate and post a comment

(39)
(8)

Raul Elizalde

Re: sytec 3000i aperture error

September 20, 2005 10:19

When you run the test from TGP what kind of error do you get?.... some times you can see two leds light on at the same time, and this is the problem, you have to identificate what leds on, and try to aligment the sensor. The problem is intermitent?....