hi to all

i have a permanant artifact on sytec 3000i images

artifact appears only with abdominal exams , all

other exams are ok

please help

thanks in advance

Log inor Register

to rate and post a comment

(107)
(6)
(31)

Milner A. Flecha

Re: sytec 3000i artifact problem

July 01, 2011 02:37

Did you check the x-ray tube is centered.?
Did you try phantom calibration. ? (large phantom)

Log inor Register

to rate and post a comment

(75)
(8)
(7)

JIFFY PAUL AKKARAKARAN

Re: sytec 3000i artifact problem

July 01, 2011 04:46

Could you describe the type of artifact coming in the system when doing the abdomen scan? is that ring or streak artifact?

Log inor Register

to rate and post a comment

(123)
(12)
(7)

Manuel Brugat

Re: sytec 3000i artifact problem

July 01, 2011 06:01

Dear Tahar:

It all depends of what type of artifact you have: Is it a pattern that repeats from the center of the image to the border with concentric rings?

If it is laike that I can almost assure you that the artifact is present in all the images. It probably is not obvious because there is not much absortion than in abdomen. I am sure that if you do a head phantom and the you window and level you finally will be able to see the same pattern.
Try to do full calibration of air and all phantoms and see if in that way the pattern go away.
I would like to know what type of image artifact you have. Is it possible you can send me a picture?

That is all I can tell you for now.

Sincerely:

Manuel Brugat/Image Instrumentation, inc

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: sytec 3000i artifact problem

July 02, 2011 12:41

Is the artifact on the outside of the image?