tahar nouri

sytec 3000i boot problem

December 24, 2011 05:28

hi to all

when i power up the CT ,on the boot sequence the

system stays with this message displayed on both

screnns " system preparition" , pls help

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

Re: sytec 3000i boot problem

December 25, 2011 11:16

There is some interlock open. Check the emergency stops buttons, the Anti-Panic tapes, the cable connections behind the front gantry cover, etc...
Good Luck!

Gus

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: sytec 3000i boot problem

December 25, 2011 11:17

Its probably a software crash.
However you could make the power on test on CPW, using the dipsw on board

Log inor Register

to rate and post a comment

Charbel Tadros

Re: sytec 3000i boot problem

December 26, 2011 10:13

need to know the led status on cpu board
so you can know where the system is stuck.

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

Re: sytec 3000i boot problem

December 26, 2011 11:26

Also, you can check the Service Status on the TGP Board, in the Left side of the Gantry.
Please, check the Service Switch is on "Normal" mode, etc, etc.
Let us know if you could solve it!
Regrads,

Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

Re: sytec 3000i boot problem

December 28, 2011 02:35

thanks to all mates
i solved the probleme , it was a misconnnected cable from the gantry . but when i make a warm-up i got error message ERR-152 , in details :
****** 2011/12/05******
*** ER-152 12:01 Warm-up error LCP=03 TN=WARM ADD.=002D3152
Error on REF_Check = REF 1 Ch
*** ER-152 12:01 Warm-up error LCP=03 TN=WARM ADD.=002D343C
Error on SD_Check = SD 1 Ch. (Active)
*** ER-152 12:01 Warm-up error LCP=03 TN=WARM ADD.=003D3A32
Illegal data on Rate_Check = Rate 1 Ch.(Active)
*** ER-152 12:01 Warm-up error LCP=03 TN=WARM ADD.=002D3DE6
Illegal Data on HPF_Check = HPF 25
when i enter in maintenace mode and do auto calibration i got error 153 with REF channel error
pls help

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

Re: sytec 3000i boot problem

December 30, 2011 09:15

OK my friend, I´m glad you can solve your problem.
Now, for the 152 Error you must know a few things.
The errors between 99 and 199 are Scan Errors.
In this case, the 152 WarmUp error, is telling you that the reference or the data collected is out of range. (the SD/MeanValue

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

Re: sytec 3000i boot problem

December 30, 2011 10:44

I´m sorry, I was wrong. I´, really sorry.
You must do the following: Start an Air Calibration, and when ask for "Initialize air data" answer YES.
Perform this 2 or 3 times and the problem should be solve.
If not, you must perform and phantom calibration.
I hope this help, and again, I´m sorry for the mistake.

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

Re: sytec 3000i boot problem

January 06, 2012 09:07

Did you fix the problem??. Please, post it here what you did, so everybody can use it for future assistance.
Thks.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)