SYTEC 3000I, WARM UP ERROR
July 07, 2009 11:47
HI
AFTER THE LAST ROTATION IN THE WARM UP, IT COME AN ERROR,
(ER-152, WARM UP ERROR, SD,)
ANY HELP,
THANKS
to rate and post a comment
Elliot Marcus
Re: SYTEC 3000I, WARM UP ERROR
July 08, 2009 10:39
Victor,
This is caused when the final warmup scan file fails the test of the Standard Deviations for each channel.
If you are familiar with the diagnostic software, go to the support menu and run the Warm Up Error Analysis to further diagnose the problem.
You can download the diagnostic service manual for the 3000i at GE Healthcare's website
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=CT