SYTEC 4000i ERROR WARM UP
June 02, 2010 12:21
HI ALL: I have a problem with SYTEC 4000i in warm up error: err: 152 WARMUP ERROR LCP.03 TN WARM ADD. 002D3072 HPF2059 Thank“s in advance
to rate and post a comment
Elliot Marcus
Re: SYTEC 4000i ERROR WARM UP
June 02, 2010 01:23
Go to your support maintenance program and perform the Warm-Up Error Analysis. The HPF (the High Pass Filter) test is failing.
Ronald Liem
June 06, 2010 10:05
Manuel, Check the DAS PS +5V, +15V, -15V located below The DAS/Detector. email me ronaldliem@yahoo.com