Hi there!
I am in trouble with this error that is generated when I try to move the column( scanning) to the foot end. Movement start and after a few milliseconds stops and generates 080/ 243 on geometry console.
Can you help me?. Any information will be appreciated!.

Log inor Register

to rate and post a comment

(13)

EDUARDO BUSTOS

re: Telediagnost error 080/243

December 09, 2020 10:19

Apparently there seems to be a problem with the safety actuators

DESCRIPTION CODE TYPE

Scanning to feet Serial OFF / Output ON 080 - 243 F

Error causing switch-off of supply to stand motors, safety switches, (K1L switched off).Error types

I Information
W Warning
E Error able to be reset
F Failed unable to be reset