Tomoscan LX Problem / IP time over

sunil Kumar

Tomoscan LX Problem / IP time over

September 21, 2006 07:33

I am getting IP time over while doing normal scan.scano and stationary scan is ok.any suggestions??

Log inor Register

to rate and post a comment

Ralph Childs

Re: Tomoscan LX Problem / IP time over

September 22, 2006 12:41

Sunil,

Your LX is losing axial rotate encoder pulses.

Scooter
scanner1@eim.ae

Log inor Register

to rate and post a comment

sunil Kumar

Re: Tomoscan LX Problem / IP time over

September 22, 2006 01:28

have already changed the rotary encorder as well as PCB on which these pulses generate view trigger

Log inor Register

to rate and post a comment

sunil Kumar

Re: Tomoscan LX Problem / IP time over

September 25, 2006 07:43

Have cheched all the scanner,rotary encorder & console PCBs but the error remains same.Is there anything else i can check??

Log inor Register

to rate and post a comment

Ralph Childs

Re: Tomoscan LX Problem / IP time over

September 27, 2006 02:50

Sunil,

You need to make sure that your rotate speed is correct too.

Log inor Register

to rate and post a comment

Gustavo Volpi

Re: Tomoscan LX Problem / IP time over

October 10, 2007 07:40

Sunil, enter on teh service program, and run the test for RAM CHECK. May be there is an error on some of the IP´s Boards.
Check, and let me know.
regards,

Gustavo