rudy

Topcon OCT 2000

October 17, 2025 10:04

My OCT start with SLD ERROR 01.
I disassembled the machine, replaced the SLD driver board with a one taken from a working machine in order to try if It works, buy i get the same error.
The error rise even if the cable (#905) connecting the grab board to the main board of the OCT (not the PC one) is disconnected ( the cable Is ok)
I searched for some burned components on PC boad, grab board and oct mainboard buy everything looks like ok.
The only thing I noted Is that the led #5 on oct mb blinks
I found that some OCT2000 have the same error buy not found any fix
Someone can help me?
Thank you

1 person liked this
Log inor Register

to rate and post a comment

(5)

jevrem doderovic

re: Topcon OCT 2000

October 20, 2025 11:40

I have the same problem. Can your machine enter service mode?

Log inor Register

to rate and post a comment

(13)
(1)

rudy

re: re: Topcon OCT 2000

October 20, 2025 02:00

How can I enter service mode?

Log inor Register

to rate and post a comment

(5)

jevrem doderovic

re: Topcon OCT 2000

October 21, 2025 10:30

Did you manage to solve the problem?

Log inor Register

to rate and post a comment

(13)
(1)

rudy

re: re: Topcon OCT 2000

October 22, 2025 08:51

Unfortunately not.
I tried several times in different ways, releasing the finger in different moments during startup but no service mode.
Is there any other way to enter service mode? i.e. using an external keyborad or mouse?

Thank you

Rudy

Log inor Register

to rate and post a comment