TOSHIBA CAPASEE ERROR CODE PC=000F00C6

Tags: Ultrasound, TOSHIBA

(32)
(10)

Selman Ahi

TOSHIBA CAPASEE ERROR CODE PC=000F00C6

May 30, 2017 10:16

Hello dear frend,

i have toshiba ssa220A with RTC problem , device is warning = ADDRESS ERROR PC=000F00C6.

I changed the battery, but again the error continues,

Do you know dip switch setting or service manual?

yasinbicerr@hotmail.com

Many Thanks,

Regards...

1 person liked this
Log inor Register

to rate and post a comment

(82)
(19)

Uldis Verdins

re: TOSHIBA CAPASEE ERROR CODE PC=000F00C6

May 31, 2017 10:44

Hi Yasin Bicer!
Your problem was writed in dotmed
https://www.dotmed.com/virtual-trade-show/category/Service-Technicians/Discussions/Forums/Toshiba-Ssa220a-Error-Code-Need-Help/5899
In five words: switch off Capasee, find dipswitch`s on CPU board (initialise), switch on machine and after Toshiba logo switch off, dipswitch to original position and them swith on and set date and time.
Good lock,
Uldis

Log inor Register

to rate and post a comment

(32)
(10)

Selman Ahi

re: re: TOSHIBA CAPASEE ERROR CODE PC=000F00C6

June 05, 2017 09:37

Hi Uldis Vernis,

Thank you for your interest.

I looked at that topic dotmed.
I turn off the device,
I pull down the switch number 5,
I turn on the device, the display shows: INITIALIZE USER REG. DATA & PATIENT DATA ARE YOU SURE (Y / N).
I press the Y key,
but the display shows: EEPROM WRITE ERROR.

I looked for a fault on the card, but I could not find a fault

Many thanks.
Best Regards...

Log inor Register

to rate and post a comment

(82)
(19)

Uldis Verdins

re: re: re: TOSHIBA CAPASEE ERROR CODE PC=000F00C6

June 06, 2017 10:28

Hi Yasin Bicer!
I`m not shore but how I remember You need to switch more one dipswitch ti initialise EEPROM from ROM. If this solution not work - problem in EEPROM (ROM to) or CPU. Tray to reseet all chip`s in socket`s.
Good lack,
Uldis

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)